GLI International F53 User manual

Rev. 1-201 Model F53 Flow Monitor/Totalizer (universal-mount 1/2 DIN)
1
OPERATING MANUAL
Model F53
Flow Monitor/Totalizer
(Universal-mount 1/2 DIN style)
Worldwide Headquarters and Sales:
GLI International, Inc.
9020 West Dean Road
Milwaukee, Wisconsin 53224
U.S.A.
Represented By:
In the interest of improving and updating its equipment, GLI reserves the right to alter specifications to equipment at any time.
A company
Viridor Instrumentation
Phone:
Fax:
E-mail:
Web:
[414] 355-3601
[414] 355-8346
info@gliint.com
www.gliint.com

Model F53 Flow Monitor/Totalizer (universal-mount 1/2 DIN) Rev. 1-201
2
This operating manual and other GLI operating manuals
are available on GLI’s web site at gliint.com. when viewed
using Adobe’s free Acrobat reader. To get this reader, link
to Adobe through GLI’s web site or visit Adobe’s web site
at adobe.com.

Rev. 1-201 Model F53 Flow Monitor/Totalizer (universal-mount 1/2 DIN)
3
IMPORTANT SAFETY INFORMATION
This analyzer is compliant with safety standards as outlined in:
FMRC Class Numbers 3600, 3611, and 3810 (U.S.A.)
CSA C22.2 No. 142 and C22.2 No. 213 (Canada)
EN 61010-1 (European Community)
Please read and observe the following:
•Opening the analyzer door exposes you to line power voltage, if present, at terminals on TB2 and TB3
inside the enclosure. This may be hazardous. Always remove line power before entering this area in
the analyzer. However, the analyzer door assembly contains only low voltage and is completely safe
to handle.
•Wiring or repairs should only be performed by qualified personnel and only to an unpowered analyzer.
•Whenever it appears that analyzer safety is questionable, disable the analyzer to ensure against any
unintended operation. For example, an unsafe condition is likely when:
1) The analyzer appears visibly damaged.
2) The analyzer fails to operate properly or provide the intended measurements.
3) The analyzer has been stored for long periods at temperatures above 158°F (70°C).
•This analyzer must be installed by personnel specially trained in accordance with relevant local codes
and instructions contained in this operating instruction manual. Observe the analyzer’s technical
specifications and input ratings. If one line of the line power mains is not neutral, use a double-pole
mains switch to disconnect the analyzer.
HELPFUL IDENTIFIERS
In addition to information on installation and operation, this instruction manual may contain
WARNINGS pertaining to user safety, CAUTIONS regarding possible instrument malfunction, and
NOTES on important, useful operating guidelines.
WARNING:
A WARNING LOOKS LIKE THIS. IT WARNS YOU OF THE POTENTIAL
FOR PERSONAL INJURY.
CAUTION:
A CAUTION LOOKS LIKE THIS. IT ALERTS YOU TO POSSIBLE
INSTRUMENT MALFUNCTION OR DAMAGE.
☞
NOTE: A note looks like this. It alerts you to important, useful operating
information.

Model F53 Flow Monitor/Totalizer (universal-mount 1/2 DIN) Rev. 1-201
4
Definition of Equipment Symbols
This symbol means CAUTION and alerts you to possible danger or
instrument malfunction. Refer to this manual before proceeding.
This symbol means that this is a protective ground terminal and
alerts you to connect an earth ground to it.
This symbol means that there is alternating current present and
alerts you to be careful.
WARRANTY
GLI International, Inc. warrants the Model F53 to be free from defects in material
or workmanship for a period of 2 years (24 months) from the date of shipment of
this product from our facility. A warranty claim will not be honored if defects are
not reported within the warranty period, or if GLI International determines that
defects or damages are due to normal wear, misapplication, lack of mainte-
nance, abuse, improper installation, alteration, or abnormal conditions. GLI
International’s obligation under this warranty shall be limited to, at its option, re-
placement or repair of this product. The product must be returned to GLI
International, freight prepaid, for examination. The product must be thoroughly
cleaned and any process chemicals removed before it will be accepted for re-
placement or repair. GLI International’s liability shall not exceed the cost of the
product. Under no circumstances will GLI International be liable for any inciden-
tal or consequential damages, whether to person or property. GLI International
will not be liable for any other loss, damage or expense of any kind, including
loss of profits, resulting from the installation, use, or inability to use this product.

Rev. 1-201 Model F53 Flow Monitor/Totalizer (universal-mount 1/2 DIN)
5
CONDENSED OPERATING INSTRUCTIONS
This manual contains details for all operating aspects of the instrument. The following condensed in-
structions are provided to assist you in getting the instrument started up and operating as quickly as
possible. These condensed instructions are only for basic operation using a GLI impeller flow
sensor. To obtain a calculated measurement, use a non-GLI flow sensor or specific features of the in-
strument, refer to the appropriate sections in this manual for instructions.
A. CONNECTING SENSOR(S)
After properly mounting the analyzer (PART TWO, Section 2), connect up to four GLI impeller flow
sensor(s), matching wire colors to terminals as indicated:
Sensor A Wire Colors Connect To Sensor C Wire Colors Connect To
Red Terminal #10 on TB1 Red Terminal #14 on TB1
Black Terminal #11 on TB1 Black Terminal #15 on TB1
Cable shield Grounding strip lug Cable shield Grounding strip lug
Sensor B Wire Colors Connect To Sensor D Wire Colors Connect To
Red Terminal #12 on TB1 Red Terminal #16 on TB1
Black Terminal #13 on TB1 Black Terminal #17 on TB1
Cable shield Grounding strip lug Cable shield Grounding strip lug
B. CONNECTING LINE POWER
Important: Follow instructions in PART TWO, Section 3.5 to connect line power to the analyzer.
C. ADJUSTING DISPLAY CONTRAST
Ambient lighting conditions may make it necessary to adjust display contrast to improve visibility.
With the MEASURE screen displayed, press and hold the ENTER key and simultaneously press the
×
×
or
Ø
Ø
key until attaining the desired contrast.
The following flow rate and volume readout instructions configure Sensor A.
Set up other sensors in the same way using their respective menu screens.
D. SETTING FLOW RATE READOUT (multiplier, decimal, and measurement unit annunciator)
Displayed flow rates are always shown with a user-set multiplier. Select measured flow rates to
be whole numbers or decimals (XXX.X or XX.XX), and choose the measurement unit annunciators.
1. Press MENU key to display
&21),*85(
7(670$,17
(;,7
.
2. Use
Ø
Ø
key to select the “CONFIGURE”line, and press ENTER key to display .
(continued on next page)

Model F53 Flow Monitor/Totalizer (universal-mount 1/2 DIN) Rev. 1-201
6
CONDENSED OPERATING INSTRUCTIONS
D. SETTING FLOW RATE READOUT -- (continued)
3. Use
Ø
Ø
key to select the “SENSOR A”line, and press ENTER key to display .
4. With the “SET MULTIPLIER”line selected, press ENTER key to display a screen like
. Use
Ø
Ø
and
×
×
keys to view the choices (x1, x10, x100, x1000, or x10000).
With the desired choice displayed press ENTER key to enter this selection.
5. After the “SENSOR A”screen re-appears, use
Ø
Ø
key to select the “SET DECIMAL”line, and
press ENTER key to display a screen like . Use
Ø
Ø
and
×
×
keys to view the
choices (XXXX, XXX.X, or XX.XX). With the desired choice displayed, press ENTER key to enter
this selection.
6. After the “SENSOR A”screen re-appears, use
Ø
Ø
key to select the “FLOW UNITS”line, and
press ENTER key to display a screen like . Use
Ø
Ø
and
×
×
keys to view all
choices. With the desired choice displayed, press ENTER key to enter this selection.
E. SETTING VOLUME READOUT (volume unit annunciator and totalizer reset mode)
Unlike flow rate, volume is shown only as a whole number without a multiplier. Choose from a
variety of volume measurement unit annunciators. The totalizer is factory-set for auto reset, which
resets the displayed volume to zero after reaching its maximum limit. After automatically resetting,
the totalizer starts a new count. Conversely, the totalizer can be set to a manual reset mode to pre-
serve the total volume. In the manual reset mode, the totalizer can be manually reset back to zero at
any time by using the “RESET VOL”function in the TEST/MAINT menu.
1. With the screen displayed, use
Ø
Ø
key to select the “VOLUME UNITS”line, and
press ENTER key to display a screen like . Use
Ø
Ø
and
×
×
keys to view all
choices. With the desired choice displayed, press ENTER key to enter this selection.
2. After the “SENSOR A”screen re-appears, use
Ø
Ø
key to select the “TOTALIZER MODE”line,
and press ENTER key to display a screen like . Use
Ø
Ø
and
×
×
keys to view
both choices (AUTO RESET or MANUAL RESET). With the desired choice displayed, press
ENTER key to enter this selection.
(continued on next page)

Rev. 1-201 Model F53 Flow Monitor/Totalizer (universal-mount 1/2 DIN)
7
CONDENSED OPERATING INSTRUCTIONS
F. SELECTING SENSOR TYPE/ENTERING SENSOR DATA (calibration)
To configure/calibrate the analyzer so that measured values will accurately represent actual values,
you must select the sensor type and enter specific sensor-related data. This is a normal part of the
initial setup. The data you enter depends on the sensor being used:
•GLI sensor mounted in tee: Enter the TEE SIZE and TEE MATERIAL.
•GLI sensor mounted in pipe (using weldolet, pipe saddle or ball valve assembly): Enter
the inside diameter of the pipe (SET PIPE ID).
•Non-GLI sensor (any mounting): Enter a slope value (SET SLOPE) and an offset value (SET
OFFSET).
After entering appropriate sensor data, the analyzer is calibrated. Anytime thereafter, you can line-
arly offset measured flow readings by using the CALIBRATE menu. Calibration adjustment should
only be considered when analyzer readings need to match a user-supplied “qualified refer-
ence” flow instrument. See PART THREE, Sections 5.1 and 5.2 for complete adjustment details.
NOTE: The following configuration/calibration procedure is only for Sensor A and only for
GLI sensors. Configure/calibrate other sensors in the same way using their respective menu
screens. (To enter specific data for non-GLI sensors, see PART THREE, Section 4.2, sub-
section “SET SENSOR Data”under the “Non-GLI Sensor Setup”category.)
1. With the screen displayed, use
Ø
Ø
key to select the “SELECT SENSOR”line,
and press ENTER key to display a screen like . Use
Ø
Ø
and
×
×
keys to select
the type of GLI sensor and mounting arrangement you are using:
•GLI PIPE MOUNT--------for sensor mounted in pipe using weldolet, pipe saddle or ball valve
•GLI POLYPROP/T -------for polypropylene sensor mounted in tee
•GLI PVDF/T ---------------for PVDF sensor mounted in tee
•GLI 316SS/T --------------for stainless steel-sleeved sensor with 2-inch NPT hex adapter mounted in tee
•GLI BRASS/T -------------for brass-sleeved sensor with 2-inch NPT hex adapter mounted in tee)
2. With the appropriate choice displayed, press ENTER key to enter this selection.
3. After the “SENSOR A”screen re-appears, use
Ø
Ø
key to select the “SET SENSOR”line. De-
pending on the selected sensor, pressing ENTER key displays:
when “GLI POLYPROP/T,”“PVDF/T,”“316SS/T”or “BRASS/T”was selected.
when “GLI PIPE MOUNT”was selected.
(continued on next page.)

Model F53 Flow Monitor/Totalizer (universal-mount 1/2 DIN) Rev. 1-201
8
CONDENSED OPERATING INSTRUCTIONS
F. SELECTING SENSOR TYPE/ENTERING SENSOR DATA -- (continued)
GLI Tee Mount Sensor Setup (select TEE SIZE and TEE MATERIAL)
A. With the “SET SENSOR”screen displayed, use
Ø
Ø
key to select the “TEE SIZE”line and
press ENTER key to display a screen like . Use
Ø
Ø
and
×
×
keys to select
the tee size in which the sensor is mounted:
•0.5 INCHES •1.25 INCHES •3.0 INCHES
•0.75 INCHES •1.5 INCHES •4.0 INCHES
•1.0 INCHES •2.0 INCHES
•1 IN. MHF15L2* •2.5 INCHES
*Select only when using a GLI PVC 1-inch MHF15L2 tee for higher flow range (up to 65 GPM).
B. With the appropriate tee size displayed, press ENTER key to enter this selection.
C. After the “SET SENSOR”screen re-appears, use
Ø
Ø
key to select the “TEE MATERIAL”line
and press ENTER key to display a screen like . Use
Ø
Ø
and
×
×
keys to
select the tee material, and press ENTER key to enter this selection. (Selected tee size de-
termines available tee materials.)
This completes GLI tee mount sensor data entry/calibration. The analyzer is now ready to meas-
ure Sensor A flow/volume.
GLI Pipe Mount Sensor Setup (SET PIPE ID)
A. Determine actual inside pipe diameter. Actually measure it or refer to an appropriate table.
NOTE: Pipe IDs for commonly-used Schedule 40 and Schedule 80 pipes are listed in Ta-
bles A and B in PART THREE, Section 4.2 under the “SET SENSOR Data”
subheading.
B. With the “SET SENSOR”screen displayed, use
Ø
Ø
key to select the “SET PIPE ID”line and
press ENTER key to display a screen like .
C. Adjust the displayed value to exactly match the actual pipe ID (inside diameter). Starting with
the extreme right blinking digit, use
Ø
Ø
and
×
×
keys to adjust its value. (The longer the key is
pressed, the faster the value changes.) Then press
Õ
Õ
key once to select the next digit, and
use the
Ø
Ø
and
×
×
keys to adjust its value. Repeat this procedure until desired value is dis-
played, and press ENTER key to enter it.
This completes GLI pipe mount sensor data entry/calibration. The analyzer is now ready to measure
Sensor A flow/volume.
G. COMPLETING ANALYZER CONFIGURATION
To further configure the analyzer to your application requirements, use the appropriate CONFIGURE
screens to make selections and “key in”values. Refer to PART THREE, Section 4 for complete con-
figuration details.

Rev. 1-201 Model F53 Flow Monitor/Totalizer (universal-mount 1/2 DIN)
9
TABLE OF CONTENTS
PART ONE - INTRODUCTION
SECTION 1 GENERAL INFORMATION
1.1 Capability Highlights ........................................................................15-17
1.2 Modular Construction ............................................................................17
1.3 Retained Configuration Values .............................................................18
1.4 Analyzer Serial Number ........................................................................18
1.5 EMI/RFI Immunity..................................................................................18
SECTION 2 SPECIFICATIONS....................................................................................19-20
PART TWO - INSTALLATION
SECTION 1 UNPACKING.................................................................................................21
SECTION 2 MECHANICAL REQUIREMENTS
2.1 Location.................................................................................................21
2.2 Mounting...........................................................................................21-23
2.3 Conduit Hole Requirements ..................................................................23
SECTION 3 ELECTRICAL CONNECTIONS
3.1 Flow Sensor(s) .................................................................................25-26
3.2 Analog Outputs.................................................................................26-27
3.3 Relay Outputs...................................................................................27-28
3.4 Pulsed Outputs (contact closure) ..........................................................28
3.5 Closed Contact TTL Input .....................................................................28
3.6 Line Power ............................................................................................29
PART THREE - OPERATION
SECTION 1 USER INTERFACE
1.1 Display .............................................................................................30-31
1.2 Keypad .............................................................................................31-32
1.3 MEASURE Screen (normal display mode) .......................................32-33
SECTION 2 MENU STRUCTURE
2.1 Displaying Main Branch Selection Screen ............................................34
2.2 Displaying Top-level Menu Screens.................................................34-35
2.3 Displaying Submenu Screens ...............................................................35
2.4 Adjusting Edit/Selection Screen Values ................................................36
2.5 Entering (Storing) Edit/Selection Screen Values/Choices.....................36

Model F53 Flow Monitor/Totalizer (universal-mount 1/2 DIN) Rev. 1-201
10
TABLE OF CONTENTS (continued)
SECTION 3 ADJUSTING DISPLAY CONTRAST.............................................................36
SECTION 4 ANALYZER CONFIGURATION
4.1 Selecting LANGUAGE to Operate Analyzer..........................................37
4.2 Configuring Sensors (A, B, C, and D):
SET MULTIPLIER (for displayed flow rate).................................38-39
SET DECIMAL (for displayed flow rate)...........................................39
SET FILTER Time.......................................................................39-40
Select FLOW UNITS (for displayed flow rate) ............................40-41
Select VOLUME UNITS (for displayed volume)...............................41
Set LO FLOW CUTOFF Limit ..........................................................42
Select TOTALIZER MODE (auto or manual reset) .....................42-43
ENTER NOTE (top line of MEASURE screen).................................43
SELECT SENSOR Type .............................................................43-44
SET SENSOR Data:
GLI Tee Mount -- Select TEE SIZE and TEE MATERIAL......44-45
GLI Pipe Mount -- SET PIPE ID..................................................46
Non-GLI Sensor -- SET SLOPE and SET OFFSET Values......47-48
4.3 Configuring CALCULATION Measurement:
SET VARIABLE X ............................................................................49
SET VARIABLE Y ............................................................................50
SET FORMULA (X/Y, X-Y, X+Y, or [X/Y] x 100) ..............................50
Select DISPLAY FORMAT (decimal and multiplier)....................51-52
4.4 Configuring Analog Outputs (1 and 2):
SET PARAMETER (representation) ...........................................53-54
SET 0/4 mA and 20 mA VALUES................................................54-55
SET FILTER Time............................................................................55
Select SCALE 0 mA/4 mA (low endpoint) ........................................55
4.5 Configuring Relays (A, B, C, and D):
SET PARAMETER (representation) ...........................................56-57
SET FUNCTION Mode (alarm, control or status).............................57
ACTIVATION (configuration values)...........................................58-59
4.6 Configuring Pulse Outputs (A and B):
SET PARAMETER (representation) ................................................60
SET VOLUME Increment ............................................................60-61
Set PULSE DURATION ...................................................................61
4.7 SET PASSCODE (feature enabled or disabled) ...................................62
4.8 Configuration Setting Summary (ranges/choices and defaults) .......63-65
SECTION 5 ANALYZER CALIBRATION
5.1 Important Information ............................................................................66
5.2 OFFSET BY or SET TO Calibration Adjustment ..............................67-68
5.3 Analog Outputs (1 and 2) Calibration...............................................68-69

Rev. 1-201 Model F53 Flow Monitor/Totalizer (universal-mount 1/2 DIN)
11
TABLE OF CONTENTS (continued)
SECTION 6 TEST/MAINTENANCE
6.1 STATUS Checking (analyzer, sensors, and relays).........................70-71
6.2 RESET VOL (A, B, C, and D) ................................................................72
6.3 HOLD OUTPUTS ..................................................................................72
6.4 OVERFEED RESET (relay timers)........................................................73
6.5 OUTPUT (1 and 2) Analog Test Signals ..........................................73-74
6.6 RELAY (A, B, C, and D) Operating Test................................................74
6.7 PULSE (A and B) Operating Test..........................................................75
6.8 EPROM VERSION Checking ................................................................75
6.9 SELECT SIM Measurement ..................................................................76
6.10 SIM Setting............................................................................................77
6.11 RESET CONFIGURE ............................................................................78
SECTION 7 RELAY OVERFEED TIMER FEATURE
7.1 Why Use an Overfeed Timer.................................................................79
7.2 Relay Overfeed Timer Setup Guidelines...............................................79
7.3 Overfeed Timer “Timeout”Operation ....................................................79
7.4 Resetting Overfeed Timers ...................................................................79
7.5 Interactions with Other Analyzer Functions......................................79-80
SECTION 8 HART OPTION
8.1 Introduction ...........................................................................................81
8.2 Analyzer Operating Modes for HART Network.................................82-83
8.3 SINGLE MODE (Point-to-Point) Wiring Arrangement ...........................83
8.4 MULTI-DROP Wiring Arrangement .......................................................84
8.5 HART Preferences Setup:
Changing Polling Address ...............................................................85
Viewing Number of Required Preambles ....................................85-86
8.6 Device Preferences Setup:
Viewing Final Assembly Number .....................................................86
Viewing Model Number ...............................................................86-87
Viewing Manufacturer ......................................................................87
Assigning a Tag ...............................................................................87
Assigning a Descriptor.....................................................................88
Assigning a Message.......................................................................88
Assigning User-defined Date ......................................................88-89
Viewing Identification (ID) ................................................................89
Viewing Revisions............................................................................89
8.7 “Master Reset”Function........................................................................90
8.8 “Refresh”Function ................................................................................90
8.9 Protocol Command Set for PC Programming ........................................90

Model F53 Flow Monitor/Totalizer (universal-mount 1/2 DIN) Rev. 1-201
12
TABLE OF CONTENTS (continued)
PART FOUR - SERVICE AND MAINTENANCE
SECTION 1 GENERAL INFORMATION
1.1 Inspecting Sensor Cables .....................................................................91
1.2 Replacing Fuse(s) .................................................................................91
1.3 Replacing Relays ..................................................................................91
SECTION 2 PRESERVING MEASUREMENT ACCURACY
2.1 Keeping Sensor(s) Clean ......................................................................92
2.2 Avoiding Electrical Interference.............................................................92
SECTION 3 TROUBLESHOOTING
3.1 Checking Electrical Connections...........................................................92
3.2 Verifying Sensor Operation ...................................................................92
3.3 Verifying Analyzer Operation ...........................................................92-93
3.4 Verifying Interconnect Cable Integrity ...................................................93
SECTION 4 ANALYZER REPAIR/RETURN
4.1 Customer Assistance.............................................................................94
4.2 Repair/Return Policy .............................................................................94
PART FIVE - SPARE PARTS AND ACCESSORIES
......................................................................................................................95

Rev. 1-201 Model F53 Flow Monitor/Totalizer (universal-mount 1/2 DIN)
13
TABLE OF CONTENTS (continued)
ILLUSTRATIONS
Figure 1-1 EMI/RFI Immunity Diagram..............................................................................................18
Figure 2-1 Analyzer Mounting Arrangements ....................................................................................22
Figure 2-2 Analyzer Installation Dimension Details............................................................................23
Figure 2-3 Analyzer Terminal Block Designations .............................................................................25
Figure 2-4 Connecting Flow Sensor(s) ..............................................................................................26
Figure 2-5 Connecting Control/Alarm Device(s) to Electromechanical Relay(s).................................28
Figure 2-6 Connecting 115 Volt Single Phase Line Power (90-130 VAC)...........................................29
Figure 2-7 Connecting 230 Volt Single Phase Line Power (180-260 VAC).........................................29
Figure 2-8 Connecting 230 Volt Split Phase Line Power (180-260 VAC)............................................29
Figure 3-1 Analyzer Keypad..............................................................................................................31
Figure 3-2 Location of SINGLE MODE/MULTI-DROP Switch (HART-equipped analyzers only) ........83
Figure 3-3 HART SINGLE MODE (Point-to-Point) Wiring Arrangement (for single analyzer).............83
Figure 3-4 HART MULTI-DROP Wiring Arrangement (for multiple analyzer network)........................84
TABLES
Table A Schedule 40 Pipe IDs.......................................................................................................46
Table B Schedule 80 Pipe IDs.......................................................................................................46
Table C Relay Configuration Settings............................................................................................58
Table D Analyzer Configuration Settings (Ranges/Choices and Defaults).................................63-65
Table E Relay Overfeed Timer Interactions with Other Analyzer Functions...................................80

Model F53 Flow Monitor/Totalizer (universal-mount 1/2 DIN) Rev. 1-201
14

PART ONE - INTRODUCTION SECTION 1 - GENERAL INFORMATION
Rev. 1-201 Model F53 Flow Monitor/Totalizer (universal-mount 1/2 DIN)
15
PART ONE - INTRODUCTION
SECTION 1
1.1 Capability Highlights
Sensor Inputs
MEASURE Screen
The analyzer has four sensor inputs to independently measure
up to four flow rates/volumes. Each input can accept:
•Any GLI IMPELLER flow sensor
•A NON-GLI SENSOR with a 0-2000 Hz. output
(paddle wheel, turbine, vortex, etc.). However, its
slope and offset values must be determined and
entered to calibrate the analyzer.
Depending on the number of sensors used, the MEASURE
screen (normal display mode) can show up to 8 separate
sensor readouts, 8 combined data readouts, and a calcu-
lated value (if configured) by pressing
Õ
Õ
or
Ö
Ö
key:
Separate Sensor Readouts
1. SENSOR A flow rate with multiplier
2. SENSOR B flow rate with multiplier
3. SENSOR C flow rate with multiplier
4. SENSOR D flow rate with multiplier
5. SENSOR A volume (no multiplier)
6. SENSOR B volume (no multiplier)
7. SENSOR C volume (no multiplier)
8. SENSOR D volume (no multiplier)
Combined Data Readouts
1. Sensor A flow rate and volume
2. Sensor B flow rate and volume
3. Sensor C flow rate and volume
4. Sensor D flow rate and volume
5. Sensor A and Sensor B flow rates
6. Sensor A and Sensor B volumes
7. Sensor C and Sensor D flow rates
8. Sensor C and Sensor D volumes
CALCULATED* Measurement Readout*
Calculation can be X-Y, X+Y, X/Y, or [X/Y] x 100 with selectable X and Y
variables representing specific sensor flow rates and volumes
*Calculated measurement can only be provided when two sensors are
used and analyzer has been correctly configured for CALCULATION.
The bottom auxiliary display line, shown in reverse video on
the MEASURE screen, can be changed by pressing the
Ø
Ø
and
×
×
keys to show analog output 1 and 2 mA values.
GENERAL INFORMATION

PART ONE - INTRODUCTION SECTION 1 - GENERAL INFORMATION
Model F53 Flow Monitor/Totalizer (universal-mount 1/2 DIN) Rev. 1-201
16
Passcode-protected
Access
Calibration
Analog Outputs
For security, you can enable a passcode feature to restrict ac-
cess to configuration and calibration settings to authorized
personnel only. See PART THREE, Section 4.7 for details.
Calibration is actually achieved by independently configur-
ing each sensor, which is a normal part of the initial setup.
First, select the type of sensor being used:
•GLI PIPE MOUNT -------for sensor mounted in pipe using weldolet, pipe
saddle or ball valve assembly
•GLI POLYPROP/T-------for polypropylene sensor mounted in tee
•GLI PVDF/T ---------------for PVDF sensor mounted in tee
•GLI 316SS/T --------------for stainless steel-sleeved sensor with 2-inch NPT
hex adapter mounted in tee
•GLI BRASS/T -------------for brass-sleeved sensor with 2-inch NPT hex
adapter mounted in tee
•NON-GLI Sensor---------for sensor with 0-2000 Hz. output (any mounting)
For GLI PIPE MOUNT, enter the internal diameter of the
pipe in which the sensor will be installed. When using a GLI
tee-mounted sensor, enter the tee size and tee material. For
a NON-GLI SENSOR, determine and enter appropriate
SLOPE and OFFSET values.
The CALIBRATE menu enables you to linearly offset meas-
ured flow readings one of two ways by using the:
•OFFSET BY method to enter a “+”or “-”offset value.
•SET TO method to enter a known measured value.
See PART THREE, Sections 5.1 and 5.2 for details. Also,
the mA values for each analog output can be calibrated
(Section 5.3).
The analyzer provides two isolated analog outputs (1 and
2). Each output can be set to be 0-20 mA or 4-20 mA, and
assigned to represent one of these measurements:
•Sensor A flow rate •Sensor B volume
•Sensor B flow rate •Sensor C volume
•Sensor C flow rate •Sensor D volume
•Sensor D flow rate •Calculated measurement*
•Sensor A volume
*An analog output can only represent the calculated measurement when two sensors
are used and the analyzer has been correctly configured for CALCULATION.
Parameter (or calculated measurement) values can be en-
tered to define the endpoints at which the minimum and
maximum analog output values are desired. For analog out-
put setup details, refer to PART THREE, Section 4.4.

PART ONE - INTRODUCTION SECTION 1 - GENERAL INFORMATION
Rev. 1-201 Model F53 Flow Monitor/Totalizer (universal-mount 1/2 DIN)
17
Relays
☞
Pulsed Outputs
1.2 Modular Construction
The analyzer has four electromechanical relays with SPDT
contacts. Each relay can be set to function as a CONTROL,
ALARM or STATUS relay. CONTROL and ALARM relays
can be assigned to be driven by one of these:
•Sensor A flow rate •Sensor B volume
•Sensor B flow rate •Sensor C volume
•Sensor C flow rate •Sensor D volume
•Sensor D flow rate •Calculated measurement*
•Sensor A volume
*A relay can only be driven by the calculated measurement when two sensors
are used and the analyzer has been correctly configured for CALCULATION.
NOTE: When a relay is set to function as a STATUS relay,
it is no longer configurable. Instead, it becomes a
dedicated system diagnostic-only alarm relay that
automatically energizes when the “WARNING
CHECK STATUS”message flashes on the
MEASURE screen. This occurs when the analyzer
detects a “fail”diagnostic condition. See PART
THREE, Section 6.1 for more details.
For relay setup details, refer to PART THREE, Section 4.5.
The analyzer provides two SPDT pulsed contact closure
outputs (A and B) for remote use such as counting or
pumping. Each pulsed output can be assigned to represent
any measured volume, and set to activate at a desired vol-
ume increment for a preset time duration. Refer to PART
THREE, Section 4.6 for setup details.
The modular construction of the analyzer simplifies field
servicing and provides electrical safety. The front door/
keypad assembly uses voltages no greater than 24 VDC,
and is completely safe to handle.
Opening the analyzer door accesses terminals inside the
enclosure for electrical connections. Line power must be
connected to specifically designated terminals on TB3.
WARNING:
REMOVE LINE POWER BEFORE NEARING THIS AREA
TO PREVENT ELECTRICAL SHOCK.

PART ONE - INTRODUCTION SECTION 1 - GENERAL INFORMATION
Model F53 Flow Monitor/Totalizer (universal-mount 1/2 DIN) Rev. 1-201
18
1.3 Retained
Configuration Values
1.4 Analyzer
Serial Number
1.5 EMI/RFI Immunity
All user-entered configuration values are retained indefi-
nitely, even if power is lost or turned off. The non-volatile
analyzer memory does not require battery backup.
A label with the analyzer model number, serial number,
build date, and other items is located on top of the enclo-
sure.
The analyzer is designed to provide protection from most
normally encountered electromagnetic interference. This
protection exceeds US standards and meets European
IEC 801-series testing for electromagnetic and radio fre-
quency emissions and susceptibility. Refer to Figure 1-1
and the specifications in Section 2.1 for more information.
FIGURE 1-1 EMI/RFI Immunity Diagram

PART ONE - INTRODUCTION SECTION 2 - SPECIFICATIONS
Rev. 1-201 Model F53 Flow Monitor/Totalizer (universal-mount 1/2 DIN)
19
SECTION 2
2.1 Operational Display....................................... Graphic dot matrix LCD, 128 x 64 pixels with
LED backlighting; 1/2 inch (13 mm) main
character height; 1/8 inch (3 mm) auxiliary
information character height; menu screens
contain up to six text lines
Measurement Selectable Ranges
Flow Rate (all sensors).......... 0-9999, 0-999.9 or 0-99.99 with selectable
flow rate units and multiplier
Volume (all sensors) ............. 0-999,999,999 with selectable volume units
Calculation (X and Y are selectable):
X-Y: Flow.......................... 0-9999, 0-999.9 or 99.99 w/selected multiplier
Volume ..................... 0-999,999,999 in preset volume units
X+Y: Flow.......................... 0-9999, 0-999.9 or 99.99 w/selected multiplier
Volume ..................... 0-999,999,999 in preset volume units
X/Y (flow or volume) .......... 0-9999, 0-999.9 or 99.99 (no units)
[X/Y] x 100......................... 0-999% pass ratio
Analog Outputs (1 and 2) ...... 0.00-20.00 mA or 4.00-20.00 mA
Ambient Conditions:
Operation.............................. -4 to +140°F (-20 to +60°C); 0-95% relative
humidity, non-condensing
Storage................................. -22 to +158°F (-30 to +70°C); 0-95% relative
humidity, non-condensing
Relays: Types/Outputs .................Four electromechanical relays; SPDT (Form C)
contacts; U.L. rated 5A 115/230 VAC, 5A @ 30
VDC resistive
Operational Mode ......... Each relay (A, B, C, and D) can be assigned
to be driven by one of these measurements:
•Sensor A flow rate •Sensor B volume
•Sensor B flow rate •Sensor C volume
•Sensor C flow rate •Sensor D volume
•Sensor D flow rate •Calculated measurement
•Sensor A volume
Function Modes:
Control .................... Settings for high/low phasing, setpoint, dead-
band, overfeed timer, off delay, and on delay
Alarm .........................Settings for low alarm point, low alarm point
deadband, high alarm point, high alarm point
deadband, off delay, and on delay
Status.........................Not configurable; relay only activates when a
sensor or analyzer “fail”diagnostic WARNING
condition exists
Indicators .........................Relay A, B, C, and D annunciators indicate
respective relay status
Sensor-to-Analyzer Distance:
GLI Impeller Sensors ............ 2000 ft. (610 m) maximum
Non-GLI Sensors .................. 300 ft. (91 m) maximum
Power Requirements .................. 90-130 VAC, 50/60 Hz. (10 VA max.) or
180-260 VAC, 50/60 Hz. (10 VA max.)
SPECIFICATIONS

PART ONE - INTRODUCTION SECTION 2 - SPECIFICATIONS
Model F53 Flow Monitor/Totalizer (universal-mount 1/2 DIN) Rev. 1-201
20
2.2 Analyzer Performance
(Electrical, Analog Outputs)
2.3 Mechanical
Calibration Adjust: Offset By ..... Enter a “+”or “-”offset value that is respec-
tively added to or subtracted from
measured flow readings for a linear offset
Set To ......... Enter measurement value (derived from
qualified reference instrument reading) to
linearly offset measured flow reading
NOTE: Calibration is actually achieved by independently configuring each
sensor, which is a normal part of the initial setup.
Analog Outputs .......................... Two isolated 0/4-20 mA outputs; each with 0.004
mA (12-bit) resolution and capability to drive up
to 600 ohm loads; each output can be assigned
to represent one of these measurements:
•Sensor A flow rate •Sensor B volume
•Sensor B flow rate •Sensor C volume
•Sensor C flow rate •Sensor D volume
•Sensor D flow rate •Calculated measurement
•Sensor A volume
NOTE: Parameter (or calculated measurement) values can be entered to
define the endpoints at which the minimum and maximum mA output
values are desired (range expand).
Pulsed Outputs .......................... Two SPDT contact closures; each pulsed
output can represent a selected measured
volume, and activate at a user-set volume
increment for a desired time duration
Communication: RS-232 ........... Enables configuration and retrieval of measured
data for one analyzer using IBM-compatible PC
and optional GLI software tool kit
HART.............. Enables configuration and retrieval of measured
data for multiple analyzers over a communica-
tion link using appropriate hand-held terminal or
data system with HART software
Memory Backup (non-volatile) .... All user settings are retained indefinitely in
memory (EEPROM)
EMI/RFI Conformance................ Exceeds US and meets European standards
for conducted and radiated emissions and
immunity; certified CE compliant for applica-
tions as specified by EN 50081-1 for
emissions and EN 50082-2 for immunity
Electrical Certifications:
General Purpose (pending) ..... UL, C-UL, FM, and CENELEC
Class 1, Division 2 (pending) .. UL, C-UL, and FM: Groups A, B, C, D, F, and G
Zone 2 (pending) .................... CENELEC: Group IIC
Accuracy.................................... ± 0.1% of span
Sensitivity .................................. ± 0.1% of span
Repeatability .............................. ± 0.05% of span
Temperature Drift ....................... Zero and Span: ± 0.02% of span per °C
Response Time .......................... 1-60 seconds to 90% of value upon step change
Enclosure................................... NEMA 4X; polycarbonate face panel, epoxy-
coated cast aluminum door and case with four
1/2 inch (13 mm) conduit holes; nylon mount-
ing bracket, and stainless steel hardware
Mounting Configurations............. Panel, surface, and pipe (horizontal and
vertical) mounting
Net Weight................................. 3.5 lbs. (1.6 kg) approximately
Table of contents
Other GLI International Measuring Instrument manuals

GLI International
GLI International GLI-98144 User manual

GLI International
GLI International GLI-93711 User manual

GLI International
GLI International C53 Instruction Manual

GLI International
GLI International GLI-9143 User manual

GLI International
GLI International P53 Instruction Manual

GLI International
GLI International Accu4 Instruction Manual

GLI International
GLI International P33 Instruction Manual