Preface
Information Paragraphs
Note:
These paragraphs provide information that provides a deeper understanding of the situation, but is not
essential to the proper completion of the instructions.
IMPORTANT: These paragraphs provide information that emphasizes instructions that are essential to proper setup
of the equipment. Failure to follow these instructions carefully may cause unreliable performance.
Safety Issues
Auxiliary Equipment
Local Safety Standards
The user must make sure that he operates all auxiliary equipment in accordance with local codes, standards,
regulations, or laws applicable to safety.
Working Area
CAUTION! This symbol indicates a risk of potential minor personal injury and/or severe damage to
the equipment, unless these instructions are followed carefully.
WARNING! This symbol indicates a risk of potential serious personal injury, unless these
instructions are followed carefully.
WARNING! It is the responsibility of the user to make sure all local, county, state and national
codes, regulations, rules and laws related to safety and safe operating conditions are
met for each installation.
WARNING! For installations in potentially hazardous areas, be sure to read the Certification and
Safety Statements document at the end of this manual before beginning the
installation.
WARNING! Auxiliary equipment may have both manual and automatic modes of operation. As
equipment can move suddenly and without warning, do not enter the work cell of this
equipment during automatic operation, and do not enter the work envelope of this
equipment during manual operation. If you do, serious injury can result.
WARNING! Make sure that power to the auxiliary equipment is turned OFF and locked out before
you perform maintenance procedures on the equipment.

Preface
iv T5 Flare Gas Transducer Installation Guide
Qualification of Personnel
Make sure that all personnel have manufacturer-approved training applicable to the auxiliary equipment.
Personal Safety Equipment
Make sure that operators and maintenance personnel have all safety equipment applicable to the auxiliary
equipment. Examples include safety glasses, protective headgear, safety shoes, etc.
Unauthorized Operation
Make sure that unauthorized personnel cannot gain access to the operation of the equipment.
Environmental Compliance
Waste Electrical and Electronic Equipment (WEEE) Directive
Panametrics is an active participant in Europe’s
Waste Electrical and Electronic Equipment
(WEEE) take-back
initiative, directive 2012/19/EU.
The equipment that you bought has required the extraction and use of natural resources for its production. It may
contain hazardous substances that could impact health and the environment.
In order to avoid the dissemination of those substances in our environment and to diminish the pressure on the
natural resources, we encourage you to use the appropriate take-back systems. Those systems will reuse or recycle
most of the materials of your end life equipment in a sound way.
The crossed-out wheeled bin symbol invites you to use those systems.
If you need more information on the collection, reuse and recycling systems, please contact your local or regional
waste administration.
Visit www.bakerhughesds.com/health-safetyand-environment-hse for take-back instructions and more information
about this initiative.

T5 Flare Gas Transducer Installation Guide v
Preface

Preface
vi T5 Flare Gas Transducer Installation Guide
[no content intended for this page]

T5 Flare Gas Transducer Installation Guide vii
Contents
Chapter 1. Installing Pipe Nozzles
1.1 Introduction . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . .1
1.2 Bias 90° Installation. . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . .1
1.2.1 Identifying and Checking the Nozzle Installation Kit Components . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . 2
1.2.2 Selecting and Marking the Pipe for Nozzle Locations . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . 4
1.2.3 Installing the First Welding Boss . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . 7
1.2.4 Installing the First Nozzle . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . 9
1.2.5 Installing the Second Welding Boss . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . 11
1.2.6 Installing the Second Nozzle . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . .13
1.2.7 Hot Tapping the Pipe. . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . 14
1.2.8 Cold Tapping the Pipe . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . 14
1.3 Tilted 45° Installation . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . 15
1.3.1 Identifying and Checking the Nozzle Installation Kit Components . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . 15
1.3.2 Selecting and Marking the First Nozzle Location . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . 16
1.3.3 Determining and Marking the Second Nozzle Location . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . 18
1.3.4 Installing the First Welding Boss. . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . .21
1.3.5 Installing the First Nozzle . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . 23
1.3.6 Installing the Second Welding Boss and Nozzle . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . 25
1.3.7 Hot Tapping the Pipe. . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . .26
1.3.8 Cold Tapping the Pipe . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . .26
Chapter 2. Installing the Isolation Valves
2.1 Bias 90° Installation (Standard or Extended Velocity Range) . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . 27
2.1.1 For 3 inch Flanges . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . 27
2.1.2 For 2 inch Flanges . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . 27
2.2 Tilted 45° Installation (Standard Velocity Range). . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . 27
2.2.1 For 3 inch Flanges . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . 27
2.2.2 For 2 inch Flanges . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . 27
2.3 Tilted 45° Installation (Extended Velocity Range) . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . .28
Chapter 3. Installing the Transducer Assemblies
3.1 Introduction . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . .31
3.2 Using the Low-Pressure Insertion Mechanism . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . .31
3.2.1 Preparing for Installation. . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . .31
3.2.2 Mounting the Insertion Mechanism. . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . .31
3.3 Mounting the Bias 90 Insertion Mechanism/Transducer Assembly . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . 36
3.4 Inserting the Bias 90 Transducer into the Pipe . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . 39
3.5 Aligning the Transducers (Standard Velocity Range). . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . 42
3.6 Aligning the Transducers (Extended Velocity Range). . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . 42
3.7 Mounting the Tilted 45 Insertion Mechanism/Transducer Assembly. . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . .46
3.8 Inserting the Tilted 45 Transducer into the Pipe. . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . 52
3.9 Connecting an XAMP . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . .54
Chapter 4. Specifications
4.1 T5 Transducer Physical Specifications. . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . 61
4.2 T5 Transducer Certifications. . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . 61

Contents
viii T5 Flare Gas Transducer Installation Guide

T5 Flare Gas Transducer Installation Guide ix
Contents

Contents
x T5 Flare Gas Transducer Installation Guide

T5 Flare Gas Transducer Installation Guide 1
Chapter 1. Installing Pipe Nozzles
Chapter 1. Installing Pipe Nozzles
1.1 Introduction
Before the T5 transducers can be installed into the pipe, you will need to install pipe nozzles. Nozzles may be installed
as part of a fabricated spoolpiece or by using the hot or cold tap process with a Panametrics Sensing
Nozzle
Installation Kit
.
IMPORTANT: This procedure only applies if you are using a Nozzle Installation Kit. If you are tapping the pipe without
using a Nozzle Installation Kit, refer to the supplied drawings in your shipment. This procedure is written
and illustrated for installations on horizontal pipes. However, the procedure is the same for vertical pipe
installations.
This section describes how to install nozzles in the following configurations:
1. Bias 90o Installation
• Standard transducer spacings of 10” and 9”, with both transducers facing straight at each other, are for
applications with gas velocities up to 100 m/s (328 ft/s). This is the
standard velocity
range.
• A shorter transducer spacing of 6.4”, with only the downstream transducer rotated 6° into the flow is for
applications with gas velocities up to 120 m/s (394 ft/s). This is the
extended velocity
range.
2. Tilted 45o Installation
• Standard transducer face to face spacings, with both transducer body axes concentric and parallel to
each other are for applications with gas velocities up to 100 m/s (328 ft/s). This is the
standard velocity
range.
• A shorter transducer face to face spacing of approximately 7.85”, with only the downstream transducer
tilted 6° into the flow, is for applications with gas velocities up to 120 m/s (394 ft/s). This is the
extended
velocity
range.
1.2 Bias 90° Installation
This procedure includes the following steps:
• Identifying and checking the nozzle installation kit components
• Selecting and marking the pipe for nozzle locations
• Installing the first welding boss
• Installing the first nozzle
• Installing the second welding boss
• Installing the second nozzle
• Tapping the pipe

Chapter 1. Installing Pipe Nozzles
2 T5 Flare Gas Transducer Installation Guide
1.2.1 Identifying and Checking the Nozzle Installation Kit Components
The
Nozzle Installation Kit
contains the materials listed below. Use
Figure 1
below to help identify each component.
• 2 Nozzles (if purchased)
• 2 Welding bosses
• 1 Jig
• 1 Alignment plate (9”, 10” or 6.4” spacing)
• 1 Spacer flange
• 1 Threaded rod (1” diameter), with washer and nut
IMPORTANT: You will need eight 5/8” studs with two nuts each, or 3/4” studs with two nuts each. The 5/8” studs are
needed for 2”-150#, 2”-300# and 3”-150# flanges. The 3/4” studs are needed for 3”-300# flanges.
After you are familiar with each component, verify that the welding bosses and alignment plate shipped are for the
required transducer spacing and the pipe size described in the following steps.
Figure 1: Components for Nozzle Installation Kit
Welding Boss
Nozzle
Jig
Spacer Flange
Alignment
Threaded Rod
with Washer and Nut
Plate

T5 Flare Gas Transducer Installation Guide 3
Chapter 1. Installing Pipe Nozzles
1.2.1 Identifying and Checking the Nozzle Installation Kit Components (cont.)
1. Check the marking on the end of the welding boss. The
pipe O.D.
and the
bias dimension
(i.e., the
transducer
spacing
) are engraved on the boss as shown below. Typically,
• For flow velocities up to 100 m/s (328 ft/s), the
10 in. spacing
is used on pipes 18 in. (450 mm) in diameter
or larger, while the
9 in. spacing
is used on pipes 16 in. (400 mm) in diameter or on larger pipes when
attenuating gases are present.
• For flow velocities up to 120 m/s (394 ft/s), the
6.4 in. spacing
is used on pipes 14 in. (356 mm) in diameter
or larger.
Note:
Bias dimension refers to the distance between the center of the nozzle location and the center line of the
pipe.
2. Check the spacing between the holes on the alignment plate, as shown below. The dimensions should
correspond to the bias/spacing dimension on the welding boss.
36 in.
5 in.
Bias Dimension (Spacing)
Pipe O.D.
3.4 in. Bias = 6.4 in. Spacing
4.5 in. Bias = 9 in. Spacing
5 in. Bias = 10 in. Spacing
12.73 in.
4.5 in. Bias (9 in. Spacing)
14.14 in.
5 in. Bias (10 in. Spacing)
9.05 in.
3.2 in. Bias (6.4 in. Spacing)
Bolt location shown
for 150# rating.

Chapter 1. Installing Pipe Nozzles
4 T5 Flare Gas Transducer Installation Guide
1.2.2 Selecting and Marking the Pipe for Nozzle Locations
1. For optimum performance, you should select a location that has at least 20 pipe diameters of straight,
undisturbed flow upstream and 10 pipe diameters of straight, undisturbed flow downstream from the point of
measurement. Undisturbed flow means avoiding sources of turbulence such as flanges, elbows and tees;
avoiding swirl; and avoiding disturbed flow profiles. Never install the flowmeter downstream of control valves,
especially butterfly valves. If you cannot find a proper location, please consult with Panametrics
Flow Application
engineering.
2. Use a center finder device to locate the center of the pipe. “EYEBALLING” IS NOT ADEQUATE FOR ACCURATE FLOW
MEASUREMENT!
3. Lightly punch two marks approximately 16 in. apart on the top of the pipe, running along the center line. The two
nozzle locations will be located between these two center punch marks.
CAUTION! Correct nozzle alignment is critical to the successful operation of the flowmeter.
Therefore, all marking, positioning and welding operations must be carried out with the
utmost attention to accuracy. Unless otherwise stated, dimensional positioning of the
nozzles must be held to a tolerance of 1/16 in. (±1.6 mm) relative to each other and with
respect to the pipe centerline. The angular tolerance must be held to 1O. All hole cutting
in process piping must be performed using hot tapping equipment.
WARNING! Be sure to adhere to all applicable safety regulations.
10D
20D
D
~ 16 in.
Top View

T5 Flare Gas Transducer Installation Guide 5
Chapter 1. Installing Pipe Nozzles
1.2.2 Selecting and Marking the Pipe for Nozzle Locations (cont.)
4. Spray the area between the two punch marks on the top of the pipe with a marking dye product. Using a metal
straight edge, scribe a line between the two punch marks.
5. Along the new scribe line, mark off a length equal to your transducer spacing (10 in., 9 in. or 6.4 in.) and scribe two
lines at least 10 in. in length and perpendicular to the scribe line on different sides of the scribe line.
6. Along each of the perpendicular scribe lines, lightly punch a mark at a distance equal to half of your transducer
spacing (5 in., 4.5 in. or 3.2 in.) from the center line, depending on the bias distance for your transducers. Keep in
mind that the distance along the arc of the pipe (L) from the pipe centerline is slightly greater than the
point-to-point distance through the pipe. These marks pinpoint the centers for the nozzles. Refer to
Table 1 on
page 7
for arc distances for the most commonly used pipe sizes. If your pipe size is not shown in
Table 1
, use the
equation in the figure on page 6 to calculate the arc distance.
Scribe Line
Top View
10", 9" or 6.4"
3", 3.5" or 4.8" 3", 3.5" or 4.8"
Top View

Chapter 1. Installing Pipe Nozzles
6 T5 Flare Gas Transducer Installation Guide
1.2.2 Selecting and Marking the Pipe for Nozzle Locations (cont.)
IMPORTANT: For a 3.2 in.bias distance, the downstream port must be located on the right side of the long scribe line
when viewed from the downstream end of the pipe (see the figure below). This convention is mandatory
for the 6.4 in. transducer spacing and is optional (not required) for a 10 in. or 9 in. spacing.
Punch Marks
SL = Arc Length
along pipe
surface
R
A
Spacing
Calculate L as follows:
L = 3.142 x R x A
180
Where A = Sin -1
R = Radius in inches
S = Bias in inches
S
R
( )
End View
Flow
Downstream Port Location
for 3.2 in. bias

T5 Flare Gas Transducer Installation Guide 7
Chapter 1. Installing Pipe Nozzles
1.2.2 Selecting and Marking the Pipe for Nozzle Locations (cont.)
7. Scribe a 6 in. long horizontal centerline through each nozzle center location point and parallel to the original
centerline.
1.2.3 Installing the First Welding Boss
1. Each welding boss has four scribe marks indicating the center of the bias.Take one welding boss and line up its
scribe marks with the horizontal and perpendicular scribe marks on the pipe as shown below. Make sure you
position the boss on the pipe so that the contoured end of the boss matches the pipe arc. That is, the shorter side
of the boss should be closer to the original scribed pipe centerline.
Table 1: Calculated Arc Distances for Common Pipe Sizes
Pipe O.D. Arc Distance
5 in. Bias
18 in. (450 mm) 5.301 in. (134.6 mm)
24 in. (600 mm) 5.157 in. (131.0 mm)
30 in. (750 mm) 5.098 in. (129.5 mm)
36 in. (900 mm) 5.067 in. (128.7 mm)
42 in. (1050 mm) 5.048 in. (128.2 mm)
4.5 in. Bias
16 in. (400 mm) 4.779 in. (121.4 mm)
18 in. (450 mm) 4.713 in. (119.7 mm)
24 in. (600 mm) 4.613 in. (117.2 mm)
30 in. (750 mm) 4.571 in. (116.1 mm)
36 in. (900 mm) 4.549 in. (115.5 mm)
42 in. (1050 mm) 4.536 in. (115.2 mm)
3.2 in. Bias
14 in. (350 mm) 3.323 in. (84.4 mm)
16 in. (400 mm) 3.292 in. (83.6 mm)
18 in. (450 mm) 3.272 in. (83.1 mm)
24 in. (600 mm) 3.239 in. (82.3 mm)
30 in. (750 mm) 3.225 in. (81.9 mm)
36 in. (900 mm) 3.217 in. (81.7 mm)
42 in. (1050 mm) 3.213 in. (81.6 mm)
Scribe Lines
Top View

Chapter 1. Installing Pipe Nozzles
8 T5 Flare Gas Transducer Installation Guide
2. Clamp the boss in place, using a pipe strap or equivalent, so that it cannot move during tack welding.
3. Check the boss alignment again, and then tack weld the boss in each of the four grooves between the boss
scribe marks. Make sure you keep the boss contour flush with the pipe contour during the entire tack welding
operation. The boss is constructed of carbon steel.
4. Check the alignment again. If the boss is misaligned by 0.02 in. (0.5 mm) or more, remove the boss, grind off the
welds and reinstall the boss.
Table of contents
Other Panametrics Transducer manuals
Popular Transducer manuals by other brands

HBM
HBM Z30 Mounting instructions

S+S Regeltechnik
S+S Regeltechnik Thor II Series Operating Instructions, Mounting & Installation

Balluff
Balluff BTL6-U101-M****-A-S4 series user guide

Clark Synthesis
Clark Synthesis TST429 installation manual

HBM
HBM P3MB Mounting instructions

S+S Regeltechnik
S+S Regeltechnik THERMASGARD TM 65 operating instructions

TR-Electronic
TR-Electronic LMP-30-300 PB manual

Simrad
Simrad 38-200 COMBI C - REV D installation manual

Simrad
Simrad 38-200 COMBI C - REV D datasheet

S+S Regeltechnik
S+S Regeltechnik PREMASGARD 711x Series Operating Instructions, Mounting & Installation

SonoSite
SonoSite L52 Series user guide

Raymarine
Raymarine HYPERVISION HV-300 installation instructions