Clamp Multimeter F205 English
You have just acquired an F205 clamp multimeter and we thank you.
For best results from your device :
read this user manual attentively,
observe the precautions for its use.
Meanings of the symbols used on the device
Danger. The operator agrees to refer to this data sheet whenever this danger
symbol is encountered.
Application or withdrawal authorized on uninsulated or bare conductors at
dangerous voltages.
9 V battery.
The CE marking indicates compliance with European directives.
Double insulation or reinforced insulation.
Selective sorting of wastes for the recycling of electrical and electronic
equipment within the European Union.
In conformity with directive DEEE 2002/96/EC: this equipment must not be
treated as household waste.
AC – Alternating current.
AC and DC – Alternating and direct current.
Earth.
Risk of electric shock.
English Clamp Multimeter F205
PRECAUTIONS FOR USE
This device complies with safety standards IEC-61010-1 and 61010-2-032 for voltages of
1000V in category III or 600 V in category IV at an altitude OF less than 2000m, indoors, with
a degree of pollution not exceeding 2.
These safety instructions are intended to ensure the safety of persons and proper operation of
the device. If the tester is used other than as specified in this data sheet, the protection
provided by the device may be impaired.
The operator and/or the responsible authority must carefully read and clearly
understand the various precautions to be taken in use.
If you use this instrument other than as specified, the protection it provides may be
compromised, thereby endangering you.
Do not use the instrument in an explosive atmosphere or in the presence of flammable
gases or fumes.
Do not use the instrument on networks of which the voltage or category exceeds those
mentioned.
Do not exceed the rated maximum voltages and currents between terminals or with
respect to earth.
Do not use the instrument if it appears to be damaged, incomplete, or not properly
closed.
Before each use, check the condition of the insulation on the leads, housing, and
accessories. Any element of which the insulation is deteriorated (even partially) must be
set aside for repair or scrapped.
Use leads and accessories rated for voltages and categories at least equal to those of
the instrument. If not, an accessory of a lower category lowers the category of the
combined Clamp + accessory to that of the accessory.
Observe the environmental conditions of use.
Do not modify the instrument and do not replace components with "equivalents".
Repairs and adjustments must be done by approved qualified personnel.
Replace the battery as soon as the symbol appears on the display unit.
Disconnect all cords before opening the battery compartment cover.
Use personal protective equipment when conditions require.
Keep your hands away from the unused terminals of the instrument.
When handling the test probes, crocodile clips, and clamp ammeters, keep your fingers
behind the physical guard.
As a safety measure, and to avoid repeated overloads on the inputs of the device, we
recommend performing configuration operations only when the device is disconnected
from all dangerous voltages.
Clamp Multimeter F205 English
1
3
4
5
6
2
7
MEASUREMENT CATEGORIES
Definitions of the measurement categories :
CAT II: Circuits directly connected to the low-voltage installation.
Example: power supply to household electrical appliances and portable tools.
CAT III: Power supply circuits in the installation of the building.
Example: distribution panel, circuit-breakers, fixed industrial machines or devices.
CAT IV: Circuits supplying the low-voltage installation of the building.
Example: power lines, meters, and protection devices.
1 PRESENTATION
Item Designation
1 Jaws with centring marks (see
connection principles)
2 Physical guard
3 Switch
4 Function keys
5 Display unit
6 Terminals
7 Trigger
English Clamp Multimeter F205
1.1 THE SWITCH
The switch has six positions. To access the , , , , , functions, set the
switch to the desired function. Each setting is confirmed by an audible signal. The functions
are described in the table below.
1.2 THE KEYS OF THE KEYPAD
1 2 3
4 5 6
Item Function
1 OFF mode – Switches the clamp multimeter off
2 AC, DC, AC+DC voltage measurement (V)
3 Continuity test
Resistance measurement Ω
Diode test
4 AC, DC, AC+DC current measurement (A)
5 Power measurements (W, var, VA) and
calculation of the power factor (PF) AC, DC,
AC+DC
6 Indicator of the order of the phases
Item Function
1 Storage of values, disabling of display
Zero correction ADC/AAC+DC/WDC/WAC+DC
Compensation of the resistance of the leads in the
continuity and ohmmeter function
2 Selection of the type of measurement (AC, DC, AC+DC)
Selection of single-phase or three-phase measurement
3 Activation or de-activation of the backlighting
of the display unit
4 Activation or de-activation of the MAX/MIN/PEAK mode
Activation or de-activation of the INRUSH mode in A
5 Measurements of frequency (Hz), of the level of harmonics
(THD)
Display of the powers W, VA, var and PF
6 Activation of the ΔREL mode – Display of relative and
differential values
1
2
3
4
5
6
Clamp Multimeter F205 English
1.3 THE DISPLAY UNIT
5
1.3.1 The symbols of the display unit
Symbol Designation
AC Alternating current or voltage
DC Direct current or voltage
AC+DC Alternating and direct current or voltage
REL Relative value, with respect to a reference
Ref Reference value
Storage of the values and hold of the display
Max Maximum RMS value
Min Minimum RMS value
Peak+ Maximum peak value
Peak- Minimum peak value
Balanced total three-phase power measurement
Item Function
1 Display of the modes selected (keys)
2 Display of the measurement value and
unit
3 Display of the MAX/MIN/PEAK modes
4 Type of measurement (AC or DC)
5 Total three-phase power
measurements
6 Display of the selected modes (switch)
7 Spent battery indication
1
2
3
4
7
6
English Clamp Multimeter F205
V Volt
Hz Hertz
W Watt
A Ampere
% Percentage
ΩOhm
m Milli- prefix
k Kilo- prefix
var Reactive power
VA Apparent power
PF Power factor
THDfTotal harmonic distortion with respect to the
fundamental
THDrTotal harmonic distortion with respect to the true RMS
value of the signal.
Indicator of order to the phases
Compensation of the resistance of the leads
Continuity test
Diode test
Permanent display (automatic switching off de-
activated)
Spent battery indicator
The display of “rdy”, for “ready”, indicates that the device is ready (“Indicator of order of the
phases” function).
The O.L (Over Load) symbol is displayed when the display capacity is exceeded.
Clamp Multimeter F205 English
1
2
1
1.4 THE TERMINALS
The terminals are used as follows:
1 2
2 USE
2.1 COMMISSIONING
Insert the battery supplied with the device as follows:
1. Using a screwdriver, unscrew the screw of the battery compartment cover (item 1)
on the back of the housing and open it.
2. Place the battery in the compartment (item 2), taking care to get the polarities right.
3. Close the battery compartment cover and screw it to the housing.
Item Function
1 Cold terminal (COM)
2 Hot terminal (+)
692875A02_Ed1_01/2012

Other manuals for F205

2

Other Chauvin Arnoux Measuring Instrument manuals

Chauvin Arnoux F27 User manual

Chauvin Arnoux

Chauvin Arnoux F27 User manual

Chauvin Arnoux DigiFlex MA4000D User manual

Chauvin Arnoux

Chauvin Arnoux DigiFlex MA4000D User manual

Chauvin Arnoux C.A 1246 User manual

Chauvin Arnoux

Chauvin Arnoux C.A 1246 User manual

Chauvin Arnoux C.A 8336 User manual

Chauvin Arnoux

Chauvin Arnoux C.A 8336 User manual

Chauvin Arnoux C.A 1550 User manual

Chauvin Arnoux

Chauvin Arnoux C.A 1550 User manual

Chauvin Arnoux C.A 1550 User manual

Chauvin Arnoux

Chauvin Arnoux C.A 1550 User manual

Chauvin Arnoux Ulys MD100 User manual

Chauvin Arnoux

Chauvin Arnoux Ulys MD100 User manual

Chauvin Arnoux C.A 1510 User manual

Chauvin Arnoux

Chauvin Arnoux C.A 1510 User manual

Chauvin Arnoux E6N User manual

Chauvin Arnoux

Chauvin Arnoux E6N User manual

Chauvin Arnoux AEMC AmpFlex 193-24-BK User manual

Chauvin Arnoux

Chauvin Arnoux AEMC AmpFlex 193-24-BK User manual

Chauvin Arnoux AEMC PowerPad 8436 User manual

Chauvin Arnoux

Chauvin Arnoux AEMC PowerPad 8436 User manual

Chauvin Arnoux C.A 6513 User manual

Chauvin Arnoux

Chauvin Arnoux C.A 6513 User manual

Chauvin Arnoux C.A 6549 User manual

Chauvin Arnoux

Chauvin Arnoux C.A 6549 User manual

Chauvin Arnoux C.A 43 User manual

Chauvin Arnoux

Chauvin Arnoux C.A 43 User manual

Chauvin Arnoux MN93 User manual

Chauvin Arnoux

Chauvin Arnoux MN93 User manual

Chauvin Arnoux pel 102 User manual

Chauvin Arnoux

Chauvin Arnoux pel 102 User manual

Chauvin Arnoux C.A 6418 User manual

Chauvin Arnoux

Chauvin Arnoux C.A 6418 User manual

Chauvin Arnoux ORITEL RW 501 User manual

Chauvin Arnoux

Chauvin Arnoux ORITEL RW 501 User manual

Chauvin Arnoux C.A 5205 User manual

Chauvin Arnoux

Chauvin Arnoux C.A 5205 User manual

Chauvin Arnoux P 310 User manual

Chauvin Arnoux

Chauvin Arnoux P 310 User manual

Chauvin Arnoux C.A 1621 User manual

Chauvin Arnoux

Chauvin Arnoux C.A 1621 User manual

Chauvin Arnoux 760N User manual

Chauvin Arnoux

Chauvin Arnoux 760N User manual

Chauvin Arnoux C.A 8230 User manual

Chauvin Arnoux

Chauvin Arnoux C.A 8230 User manual

Chauvin Arnoux pel 102 User manual

Chauvin Arnoux

Chauvin Arnoux pel 102 User manual

Popular Measuring Instrument manuals by other brands

Larson Davis 831 user manual

Larson Davis

Larson Davis 831 user manual

S&S Northern Merlin 1000BH user guide

S&S Northern

S&S Northern Merlin 1000BH user guide

Environmental Express ION 100 instruction manual

Environmental Express

Environmental Express ION 100 instruction manual

Kane 358 manual

Kane

Kane 358 manual

Phynix DC-2000C instruction manual

Phynix

Phynix DC-2000C instruction manual

Proline Promag L 400 technical information

Proline

Proline Promag L 400 technical information

NDT V-E-400 Operator's manual

NDT

NDT V-E-400 Operator's manual

Meco 4500 user manual

Meco

Meco 4500 user manual

Magtrol 7500 SERIES user manual

Magtrol

Magtrol 7500 SERIES user manual

Nexus 1262 quick start

Nexus

Nexus 1262 quick start

Alentec & Orion 24700 user manual

Alentec & Orion

Alentec & Orion 24700 user manual

X-Rite RM61 user guide

X-Rite

X-Rite RM61 user guide

Tektronix AM503S user manual

Tektronix

Tektronix AM503S user manual

KROHNE OPTIMASS Series Handbook

KROHNE

KROHNE OPTIMASS Series Handbook

Seneca R203 installation manual

Seneca

Seneca R203 installation manual

Hornady 95907 quick start guide

Hornady

Hornady 95907 quick start guide

Apera PC60-Z user manual

Apera

Apera PC60-Z user manual

Tegam 710A Operation manual

Tegam

Tegam 710A Operation manual

manuals.online logo
manuals.online logoBrands
  • About & Mission
  • Contact us
  • Privacy Policy
  • Terms and Conditions

Copyright 2025 Manuals.Online. All Rights Reserved.